Look OK to me. Did do try calibrating the probes when in X10 mode? There should be a screw on the probe shaft or there abouts that adjusts the capacitance on the tip of the probe. Idea is you adjust it till you have as close to a square wave as possible.
On a X1 probe, the calibration has no effect. So you do see that distortion of the waveform.